Answers to Is there, or has there ever been anything remotely like an electron microscope in space? show that there has indeed been one (as do answer(s) to Microscopes in space?), but mostly electron microscopy of space samples comes from either meteorites on Earth or samples returned by collection spacecraft which sometimes included humans.

Electron microscopes are not the only instruments with nanoscale resolution. Scanning probe microscopes or SPMs raster an actual physical tip over a specimen's surface in order to generate an image, rather than an electron beam.

I've just seen in this answer that Rosetta had an atomic force microscope (AFM) aboard, and it seems it was able to capture particles, bring them on board and measure them, not to mention the feat of blasting such a delicate instrument off the Earth and expose it to a space environment for years first.

Question: How did ESA blast an atomic force microscope off of the Earth and deploy it in deep space safely and reliably, then capture tiny particles from a comet and position them "under" it?

$\endgroup$
5
  • 1
    $\begingroup$ The wikipedia page for the instrument (en.wikipedia.org/wiki/Micro-Imaging_Dust_Analysis_System) seems to cover your questions, are there any specifics that are not answered? I can elaborate more / translate if that is what is desired. $\endgroup$
    – IronEagle
    Commented Nov 6, 2020 at 16:39
  • 1
    $\begingroup$ @IronEagle I'm quite sure there is much better and more detailed technical material than those few spartan sentences on how all this was done. $\endgroup$
    – uhoh
    Commented Nov 6, 2020 at 16:58
  • 2
    $\begingroup$ Alright, I’ll get started. Might take a while... $\endgroup$
    – IronEagle
    Commented Nov 6, 2020 at 16:59
  • $\begingroup$ Super! I used to do AFM myself back in the 1990's, and at least at that time it was quite delicate and very intensive hands-on work, and there were regular tip changes, especially when samples had some topography and you'd bang it into something. These are randomly shaped objects and presumably don't even have flat surfaces to scan. It is just so counterintuitive to me that this can be so robust and simultaneously automated from sample collection all the way to positioning directly under the tip, it boggles the mind! $\endgroup$
    – uhoh
    Commented Nov 6, 2020 at 17:02
  • 1
    $\begingroup$ Instrument paper and MIDAS results paper both freely available/open access $\endgroup$ Commented Nov 6, 2020 at 17:04

1 Answer 1

13
$\begingroup$

What is an atomic force microscope (AFM) in the first place?
An AFM uses a very sharp tip to "probe" around the surface of an object. As it gets closer to or further away from the surface, tiny changes in force are observed from van der Waals or similar "atomic" forces. A computer can stitch together a grid of individual passes over the surface to give an overall image. The size of the tip can allow imaging in the 0.1 - 1 nanometer range. On MIDAS, the probes were mounted to the moving microscope stage through a piezo crystal, which allowed both movement through an applied voltage, and measuring of the tip force through an induced voltage.
Sample AFM Probe (1)
Sample AFM Probe. Note the scale at the bottom of the image (1)

How did they design the AFM to survive space?
As @uhoh has commented, Earth-based AFMs are not known for their robustness, although they are generally smaller machines compared to other methods, such as Scanning Electron Microscopy. This is because they rely on an extremely sharp tip , which can be damaged through actual contact with a surface rather than just floating near it. The safety of the tip was obviously a concern, both from launch stresses and operation. For operation, 16 tips were included, with the ability to switch out the tips automatically (2). Safety during launch was fairly simple, consisting of just locking everything in place where the tips couldn't come into contact with anything. Vibrations of the spacecraft during AFM operation were handled by a mechanical damping system, which nullified vibrations with frequencies above ~5 Hz (2). Additional resiliency was provided by using the same piezo crystal for vibration and for measurement, reducing the AFM complexity and avoiding the need for separate measurement equipment close to the tip.

How did they capture and position dust (from the comet, and in general)?
The capturing and imaging was a multi-step process. First, a "horn" on the outside of the spacecraft was used to capture and guide dust towards a "shutter". This shutter opened to allow an "exposure" of one of 64 sections on a wheel, which had been specially coated with solgel. The 14 nm-thick layer of solgel allowed particles to slightly embed themselves in the surface upon impact, resulting in a overall very flat, slightly "sticky" surface.

Calibration after launch was performed by imaging an unexposed section of the wheel, which gave a very flat reference surface. Any damage to the probes could thus be detected and corrected for. After an exposure, the wheel could be rotated and then translated up to 30mm sideways to select the correct tip for use. The microscope then slowly brings the tips down to the correct height for imaging, and rasters them across the surface.
(3, see this image) enter image description here Image of MIDAS: The horn is on the left, the sample wheel is in the middle, and I believe the actual microscope tips are part of the black and white bar connected to the brown strip labeled "F15". Image courtesy Mark Bentley @ Institut für Weltraumforschung

If you have further interest, I can strongly recommend the paper found by @astrosnapper, (2) which goes into greater detail about the instrument itself. Interestingly, the whole instrument was only 8kg, and consumed only 7.4W (4).

It is also of note that this is not the only AFM in space, the Mars Phoenix lander also carried an AFM, which actually went into operation before MIDAS, due to the length of time needed to reach the comet vs. Mars. (link)

$\endgroup$
2
  • $\begingroup$ Beautiful answer thank you very much! AFM's can't really search large areas for something interesting; I'm guessing the area of the solgel if express in square nanometers was on the order of 10${}^{15}$, so I'm thinking that they must have had a parallel optical microscope to hunt for particles first, and a mechanical system that would then translate the particle found optically a known offset, positioning it under the AFM. The "horn" translated the macro world of space down to a postage stamp, but I think an optical microscope must have been used to translate further to nanoscale world. $\endgroup$
    – uhoh
    Commented Nov 7, 2020 at 0:43
  • $\begingroup$ That is a lot of big power resistors. $\endgroup$
    – ikrase
    Commented Mar 18 at 6:08